What is the purpose of a spike recovery test, and how is it interpreted?

Enhance your skills with the CWEA Grade 2 Lab Analyst Test. Study with flashcards and multiple choice questions, each offering hints and explanations. Prepare successfully for your exam!

Multiple Choice

What is the purpose of a spike recovery test, and how is it interpreted?

Explanation:
Spike recovery testing checks how accurate a method is in a real sample matrix by adding a known amount of analyte and seeing how much is recovered after sample preparation and analysis. You compute percent recovery as (measured amount) ÷ (amount spiked) × 100. Recoveries near 100% indicate the method is accurate in that matrix, while recoveries in the 80–120% range are generally considered acceptable, acknowledging that some matrix effects or small procedural losses are normal. If recoveries deviate significantly from this range, it suggests the matrix is influencing the measurement or the extraction/detection steps are not performing properly, signaling potential bias in the results. This approach focuses on method accuracy within the matrix, rather than instrument drift, detection limit determination, or merely calibrating with a matrix blank.

Spike recovery testing checks how accurate a method is in a real sample matrix by adding a known amount of analyte and seeing how much is recovered after sample preparation and analysis. You compute percent recovery as (measured amount) ÷ (amount spiked) × 100. Recoveries near 100% indicate the method is accurate in that matrix, while recoveries in the 80–120% range are generally considered acceptable, acknowledging that some matrix effects or small procedural losses are normal. If recoveries deviate significantly from this range, it suggests the matrix is influencing the measurement or the extraction/detection steps are not performing properly, signaling potential bias in the results. This approach focuses on method accuracy within the matrix, rather than instrument drift, detection limit determination, or merely calibrating with a matrix blank.

Subscribe

Get the latest from Passetra

You can unsubscribe at any time. Read our privacy policy